Question: Bug : Impossible to calibrate this mercury vapour spectrum

SarahC is asking a question about spectral-workbench
Follow this topic

by SarahC | June 06, 2018 04:11 | #16434


Where can I report this?

The blue bar appears to need to be negative, and the calibration is therefore not stored.https://spectralworkbench.org/spectrums/136374

Is there a way of adding a calibration with user-defined calibration positions?

404.7 - H-line - violet

435.8 - G-line - blue

546.1 - green

578.2 - yellow-orange



2 Comments

I found there's a manual calibration URL I could use:

https://spectralworkbench.org/spectrums/calibrate/136374?x1=165&w1=546&x2=189&w2=579

This fixed the issue for me.

(The X position of the bars is simply found by hovering over the line chart when it's uncalibrated - which gives the X value from the leftmost end of the spectral image.)